If a an alarm event is not confirmed on the device and the device gets back within the set temperature limits, an alarm will be retriggered when the limit is violated again.
This ensures that a forgotten confirmation of an alarm event on the device itself, does not result in a missed following alarm.
Events on not confirmed Alarms
Settings: upper temperature limit >8.0°C and duration >1 hour, lower temperature limit <–0.5°C and duration 1 hour
1. Alarm triggered: alarm symbol and warning symbol
on display
2. Alarm not confirmed within the set temperature limits: alarm symbol and warning symbol stay
on display.
3. Alarm retriggered: A new alarm notification is sent to the predefined recipients, the alarm symbol and warning symbol
on display.
Note: Only the first alarm on a day is recorded in the ASCII/PDF report