Power and Electrical

The following is required from the customer:

SR006 CE UL/CSA
North America
UL/CSA
North America
MODEL -05, -08, -09, -11 -06, -10 -07, -12
Power 1 PH, 230VAC, 20A 1 PH, 240VAC, 20A 1 PH, 120VAC, 20A
Receptacle Wall Mounted:
ABB
ABB332R6W
NEMA L6-20R (L6, 20A, 240 VAC, 2-POLE, 3-WIRE) NEMA L5-20R
(L5, 20A, 120VAC, 2-POLE, 3-WIRE)

Facility and Floor Requirements

Operating Environment Requirements:

  • Temperature: 5°C to 40°C (41°F to 104 °F)
  • Relative humidity: 0% to 80%.
  • Maximum altitude: 10,000 feet (3,048 meters)
  • Pollution Degree 2
  • Overvoltage Category II
  • Sufficiently sized doorways to transport the machine from loading dock to permanent installation location.
  • Smooth, level flooring. The DEMI 800 Series must be placed on a level floor.
  • Sufficient space in installation location for DEMI 800 Series and operator to maneuver around the machine and access master disconnect switch during normal operation.
  • Locate and operate the unit at least 6 inches (0.15m) from any wall from all enclosure sides
  • Installation area should be spacious and well ventilated.
  • Power disconnects, plumbing, and emergency equipment should be installed in the same room as the DEMI 800 Series.

Ventilation

PostProcess Technologies detergents are specifically formulated for maximized 3D printed support removal efficiency. Specific ventilation requirements can vary widely due to the local, state, and federal regulations, size of the install room, and existing ventilation (positive or negative).

To meet safety and environmental requirements, it is recommended that the post printing operations using PostProcess equipment and chemistry solutions have ventilation that is exhausted externally. Our VVD and TAF solutions come equipped with ventilation systems (blowers) upon request. Our SVC solutions should be located in a well ventilated area and have a ventilation draw on the open top tank. Make up air should be provided from the facility where the equipment is located.

It is recommended that customers performing post printing operations have or consult with a responsible environmental health and safety person (such as an Industrial Hygienist) to determine and verify meeting environmental, health, and safety (EH&S) requirements.

PostProcess can provide the recommended ventilation requirements for each of our solutions as well as recommendations for ventilation consultation and EH&S / Industrial Hygienists.

For more information on the safe use of PostProcess detergents refer to the SDS for the specific detergent used.

Feedback

Was this helpful?

Yes No
You indicated this topic was not helpful to you ...
Could you please leave a comment telling us why? Thank you!
Thanks for your feedback.

Post your comment on this topic.

Please do not use this for support questions.
Your feedback will remain private. Click here to Request Technical Support.

Post Comment