Power and Electrical
The following is required from the customer:
SF003 | UL/CSA | CE |
---|---|---|
MODEL | -04 & -05 | -06 |
Phases, Hz | 1, 60 Hz | 1, 50 Hz |
Voltage | 120 V | 230V |
Amperage | 15A | 16A |
Receptacle | NEMA 5-15R, 3 prong outlet | CEE 7/3, 2P+PE outlet |
Plumbing and Drainage
- DRAIN outlet: Floor Drain or Waste Container*
Facility and Floor Requirements
Operating Environment Requirements:
- Temperature: 5°C to 40°C (41°F to 104 °F)
- Relative humidity: 0% to 80%.
- Maximum altitude: 10,000 feet (3,048 meters)
- Pollution Degree 2
- Overvoltage Category II
- Sufficiently sized doorways to transport the machine from loading dock to permanent installation location.
- Smooth, level flooring. The RADOR must be placed on a level floor.
- Sufficient space in installation location for RADOR and operator to maneuver around machine during normal operation.
- Locate and operate the unit at least 6 inches (0.15m) from any wall from all enclosure sides
- Installation area should be spacious and well ventilated.
- Power disconnects, plumbing, air, and emergency equipment should be installed in the same room as the RADOR.
Ventilation
PostProcess Technologies detergents are a specifically formulated chemistry for maximized 3D printed support removal efficiency. It is recommended that the equipment be located in a well-ventilated room. Specific ventilation requirements can vary widely due to size of the install room, existing ventilation (positive or negative) and differences in individual sensitivities to airborne scents or fragrances.
For more information on the safe use of PostProcess detergents refer to the SDS for the specific detergent used.