TYPE DESCRIPTION HOW TO RESOLVE
WARNING Tank Not Full Add detergent to processing tank using the DOSE function explain in Manual Operations.
WARNING Auto PSI Low Spray valve pressure begins to fall below the desired pressure setting. This is an indication that the drain and filters need to be cleaned. See Filter Maintenance
WARNING Auto PSI High Spray valve pressure begins to rise above the desired pressure setting. This is an indication that the spray valves need to be cleaned. ▶
WARNING Low pH Level The pH balance in the tank has gone below the desired pH setting. It is typically cleared by running a dosing cycle.
WARNING Temperature Limit Exceeded The temperature has gone over the setpoint limit. While active, no device that adds heat to the system can be turned on. This warning will clear once the tank temperature falls below the temperature limit.
TYPE DESCRIPTION HOW TO RESOLVE
ALARM Low Tank Level See Low Tank Level for more detail.
ALARM Emergency Stop Confirm operating conditions are safe. Pull out e-stop button to reset.
ALARM Over Temperature See Over Temperature for more detail.
ALARM Chamber Door Open See Chamber Door Open for more detail.
ALARM Filter Screen Maintenance Due The Remaining Optimal Usage Gauge has expired. See Maintenance Mode Settings for more detail.
ALARM pH Sensor Maintenance Due The Remaining Optimal Usage Gauge has expired. See Maintenance Mode Settings for more detail.
ALARM Dosatron Maintenance Due The Remaining Optimal Usage Gauge has expired. See Maintenance Mode Settings for more detail.
ALARM PSI Extreme High Valve pressure has risen above the high tolerance set point. Check the spray valves for residue buildup. See Troubleshooting.
ALARM PSI Extreme Low Valve pressure has fallen below the low tolerance set point. Check the tank drain and screen for buildup. See Troubleshooting.
ALARM Dosing Error See Dosing Error for more detail.
ALARM Motion Error See Motion Error for more detail.
ALARM Defoamer Required See Defoamer Required for more detail.
ALARM HMI Comm Error The controller touchscreen or PLC stopped communicating with one another. To resolve, a system restart may be necessary.

Reaktion

War das hilfreich?

Ja Nein
You indicated this topic was not helpful to you ...
Could you please leave a comment telling us why? Thank you!
Thanks for your feedback.

Hinterlasse einen Kommentar.

Bitte stellen Sie hier keine Supportanfragen.
Your feedback will remain private. Click here to Request Technical Support.

Kommentar senden