Diverse information can be read out after the INFO button has been pressed 3 times in one second. After 1 minute the device goes back into sleep mode; the display is blank again. (The date format is dd/mm/yyyy.)

1st display: all segments activated
2nd display: date and time of production test results
3rd display: start delay time
4th display: current date, time and ambient temperature
5th display: temperature, time limits and ALARM type
S: single event
HI: high limit
Time: h 01 00 MIN
Explanation: Alarm occurs if the limit of 44.9°C is exceeded for 1 hour continuously.
6th display: * temperature, time limits and ALARM type
C: cumulative event
HI: high limit
Time: h 10 00 MIN
Explanation: Alarm occurs if the limit of 29.9°C is exceeded 10 hours in total (sum of Hi Range violation).
7th display: * temperature, time limits and ALARM type
S: single event
LO: low limit –0.4°C
Time: h 01 00 MIN
Explanation: Alarm occurs if the limit of –0.4°C is undercut by 1 hour continuously.
8th display: * temperature, time limits and ALARM type
C: cumulative event
LO: low limit –0.4°C
Time: h 01 40 MIN
Explanation: Alarm occurs if the limit of –1.4°C is undercut by 1 h 40 min. (sum of LO range violation).
9th display: * temperature, time limits and ALARM type
S: single event
HI: high limit
Time: h 00 01 MIN
Explanation: Alarm occurs if the limit of 8.9°C exceeds 1 minute continuously.
10th display: time expiry (max. running time in days)
Explanation: TIME EXP appears on the display after 20 days of operation.

*(only indicated if factory-preset, otherwise skipped)